Hi,
Have you tried configuring the device to use the 4Hz filter? The filters of this device are used to eliminate noise introduced by the isolation amplifier. The
manual for the 1122 also indicates the following on page 4-4:
"To achieve 50 or 60 Hz rejection, you should acquire data over an integral number of power line cycles, then average this data."
Regards,
Andrew W
National Instruments