1122 -1322 to aquire 0-1.5v input. my readings are not accurate

 I am using a 1122 scxi board with a 1322 terminal block to read voltage from several RO105 temperature sensors.  i see a +- .080v 50hz osc on the voltage reading.  Any suggestion ?
0 Kudos
Message 1 of 4
(3,487 Views)
50Hz, sounds like you are picking up noise from the AC power mains. In your location is the power line frequency 50Hz?
~~~~~~~~~~~~~~~~~~~~~~~~~~
"It’s the questions that drive us.”
~~~~~~~~~~~~~~~~~~~~~~~~~~
0 Kudos
Message 2 of 4
(3,484 Views)
that was my first thought, I connected everything thru a isolation trans an all my signals re shielded
0 Kudos
Message 3 of 4
(3,474 Views)
Hi,

Have you tried configuring the device to use the 4Hz filter?  The filters of this device are used to eliminate noise introduced by the isolation amplifier.  The manual for the 1122 also indicates the following on page 4-4:
"To achieve 50 or 60 Hz rejection, you should acquire data over an integral number of power line cycles, then average this data."

Regards,
Andrew W