05-09-2016 07:43 AM
Dear All,
I have problem understanding specification of my PXI 4072 card. On the specification data sheet its written accuracy is DC Current +/- (PPM of reading + PPM of range) , what does this PPM of reading means ?
Best Regards
Solved! Go to Solution.
05-10-2016 01:45 AM
Hi,
I would like to say: 'Why use it if you don't understand it' But I don't know your situation.
How I understand it: For DC current as in your picture the tolerance 2 years after calibration is 500ppm of the reading + 50ppm of the range for the 1A range.
So if your are measuring 500mA the tolerance is (500ppm of 500mA) 250 µA + (50ppm of 1A) 50 µA = 300 µA
Kees
05-10-2016 02:21 AM
need that parameter to calculate the measurement uncertanity of the card. Thanks