PXI
Temperature Measurements Increasing after Starting SCXI-1102/1303 Module
Hello Forum,
Hopefully this is a simple one: I have 9 type-K TCs connected to a SCXI-1303 temperature module in channels 0 - 8. My temperature readings increase to about 90 degF (in a room with no heat input, etc) over a 30 minute period in channels 0 - 7. The thermocouple in channel 8 reads 75 degF for the entire time. This is rather confusing since everything is at ambient conditions -  around 72 degF, etc. I was wondering if this might be an issue with the bias and/or pullup resistor networks for channels 0 - 7 that cause the TC readings to increase over time. Any help is appreciated!
(3,632 Views)
Accepted by topic author JamesW3157
There is no immediate answer to your question that I'm aware of.
I would try the working thermocouple in other channels before writing it off as incorrect.
Also, check out this article http://zone.ni.com/devzone/cda/tut/p/id/3461#toc1.
There are a few set up necessities with the 1102 and a 1303 such as disable the ground-referencing circuitry on the SCXI-1303 to prevent ground loops.
Daniel
(3,622 Views)
Thanks Daniel!
The article helped me diagnose what was going on.
(3,587 Views)
