PXIe-6365 accuracy

Evaluating NI 6365, on p. 3 of https://www.ni.com/pdf/manuals/374654e.pdf I observe 1.66mV accuracy in the example that assumes 10,000 measurements to reduce noise impact if I only change the number of measurements from 10,000 to one, I get ~2.6 mV, my accuracy is ~13 bits (20V range +/- 10V and the 2.6 mV absolute accuracy), does that stack-up?

 

We would like to measure +/-140V time-variant (3MHz peak frequency content) signal with 15-bit accuracy, so expect to precision voltage divider and a good trigger and then our source resistance & settle time becomes a concern!

 

Appreciate your confirmation and suggestions regarding our application.

0 Kudos
Message 1 of 3
(1,051 Views)

Try to find a calibration lab to qualify that 😄

15bit is about 30ppm, I'm not aware of any gear that can do that, but also interested in solutions 😉

 

My guess: be happy if you reach 0.5%

Here you can find the spec of a Fluke 5790 AC voltage transfer standard (40k+ gear):

http://download.flukecal.com/pub/literature/5790B___omeng0000_0.pdf

Greetings from Germany
Henrik

LV since v3.1

“ground” is a convenient fantasy

'˙˙˙˙uıɐƃɐ lɐıp puɐ °06 ǝuoɥd ɹnoʎ uɹnʇ ǝsɐǝld 'ʎɹɐuıƃɐɯı sı pǝlɐıp ǝʌɐɥ noʎ ɹǝqɯnu ǝɥʇ'


0 Kudos
Message 2 of 3
(982 Views)

Additional remarks:

The 6365 is a multiplexed card .. with 2MSPS max for one channel, 60ppm INL , >166ppm accuracy (I guess for DC one channel 😉 )  not quite the rigth gear for that task.

 

So, do you need use that card (and have the 'accademic' wish)?

or do you need these specs (and have resources) ?

 

Greetings from Germany
Henrik