Max wont let me add a temperature channel for 4351

Solved!
Go to solution

I jsut inherited a 1065 chassis with two 6363 and two 4351.  I am trying to use one of the 4351 for TC measurements but, when I try to configure one of the 4351 in MAX for a new temperature task it brings up the 6363 and not the 4351.  I must be missing somthing.

 

Please advise

0 Kudos
Message 1 of 4
(5,196 Views)

Teslac,

 

My first thought is that the names in MAX may be misleading.  Double check that the names correspond the devices they describe.  If that checks out I would power the system down, move the cards around and determine if this happens again. 

 

After that start looking into clearing the MAX configuration.  Be aware: THIS WILL REMOVE CUSTOM SCALES, DEVICE CONFIGURATION etc!!

http://digital.ni.com/public.nsf/allkb/2C7480E856987FFF862573AE005AB0D9

 

0 Kudos
Message 2 of 4
(5,172 Views)

Thanks, I will give that a try by moving the cards.

0 Kudos
Message 3 of 4
(5,142 Views)
Solution
Accepted by topic author teslac

You can not create DAQmx tasks, channels or scales in MAX for the 4351.  It is not supported by NI-DAQmx.

 

http://digital.ni.com/public.nsf/allkb/BE27E48645E2DF1386256FBC007157AC

 

You will need to use the Traditional NI-DAQ Virtual Channels with the 4351.

 

http://www.ni.com/white-paper/3445/en/