Hi, All,
in electrical measurement, like what we usually practice: output voltage and monitor current. there are two parameters related to current measurement: "current measurement range" and "current compliance". these two parameters should be separated and not correlated, in other words, if I measure a 1nA signal, I can set the compliance as high as 1000A, the system should still be able to give the same reading as the compliance set at 1uA as long as you set the "current measurement range" as auto. therefore the compliance shouldn't t affect the measurement readings except it is set too small. but in NI's case, I think the NI driver that was supposed to set the system into "auto range" is somehow messed up, it actually set the current measurement range as " measurement range >= the compliance". this is what several of us discussed and find out. if you have a NI SMU at hand, and with this example vi from NI-- "NI DCPower hardware-timed voltage sweep", you can check youself. Just measure the background noise with different "compliance" input like 10-8 vs. 10-3, you will notice the change in the readings. one thing to clear, in this vi, the current compliance is called "current limit".
at this moment, it seems NI don't have a functional driver/sub-vi to auto-set the current measurement range. then when users measure device other than a resistor, they will not succeed because you can not use one current measurement range to deal with all the data points, especially when the readings spread over several magnitude. for example if the device's current is 1nA at 1V and 2mA at 2V, then 1nA again at 3V. if you set the current range as mA, then the system's resolution would not be good enough to tell the nA. if you set the current range as uA, which should have enough resolution to tell nA, then it can not handle the mA readings. I wonder if other user have encountered the same situation or NI is aware of the situation.
regards,
ted
04-24-2019 11:01 AM
The NI DCPower Driver has several flaws that I have seen in my application, with the one you've described being the most limiting.
Technically, all 3 of these issues can be resolved by moving functionality into the software but this completely eliminates use of deterministic hardware-timed sequence or even software-timed sequences. It becomes non-uniform software-timed single points.
04-26-2019 01:32 PM
For future reference, this discussion would be better placed on the DC Power page. This way it is more likely for the correct R&D team to see your feedback.
I want to thank you for letting us know about these requests as we are always looking to improve our drivers. I also wanted to let you know I have already passed/pumped your requests to our DC Power Software Developer team. Please let me know if you have any other feedback.
Thanks!
04-26-2019 04:44 PM
Thanks for the DC Power page link. I wasn't finding the correct location and stumbled upon this fairly recent post discussing the same issue.
When sweeping a diode from 0V to 1V, I measure less than 100pA at <0.7V and up to 100mA at >0.7V. To get around this, I have to run two sweeps with different compliance ranges based on the anticipated measurement. Then, I have to attempt to concatenate them despite having different noise/resolution in the differing ranges. Finally, not all measurements would have this 'anticipated' range.
The auto-range would require additional measurement time (making measurement time less predictable) but is desirable over having to run multiple sweeps or lose resolution altogether. Changing any of these settings on-the-fly or aborting/initializing moves the measurement into software, whereas these high-end SMUs heavily advertise the hardware-sequences.
Compliance behavior can also be accomplished in software by trading off the deterministic hardware sequenced performance.
Good to know that this is on the team's radar. I have a lot of experience using traditional semiconductor parameter analyzer equipment and I'm happy to help bring these SMU's up to full speed so that I can advance my utilization of them in ATE.
Thank you for taking the time to read, respond and forward to the DC Power Software Developer team.
01-15-2021 10:18 AM
I recently bought a PXIe-4163 SMU module and have the same issue with diode measurement (-2V to +2V range, <0V in nA range and >0V in 100mA range)
the issue with dcpower driver is there's mixed up in source limit/range and measurement range.
there's no place for me to setup the measurement range.
there's an option to setup autorange for measurement, however, every time when I try to set it to on state, it gives me error. it defaults at off state and no way to change it.
01-18-2021 06:34 PM - edited 01-18-2021 06:36 PM
Providing the same reply, in NI SMU series, the measurement range and compliance ranges are one and the same because the measurement is used to implement the compliance (probably at the software level in the control system implemented on the FPGA).
As seen in the below block diagram of the 4163, the FPGA implements both the voltage and current compliance and hence there are no separate ranges for measurement and compliance.
This is something different from other vendors where the compliance may be implemented at a hardware layer by analog crowbar circuit.
01-18-2021 06:57 PM
Santhosh,
thanks for the clarification.
but then any idea how to setup autorange?
Assuming source voltage and measure current, the current range from nA to mA and I'd like to setup for autorange.(diode IV measurement)
this method doesn't work
dcPowerSession.Outputs[ChannelName].Measurement.Autorange = DCPowerMeasurementAutorange.On
I can only set it to Off state. otherwise receive an error message.
I also tried this
dcPowerSession.Outputs[ChannelName].Source.Voltage.CurrentLimitAutorange = DCPowerSourceCurrentLimitAutorange.On;
although I didn't receive an error message but I think its still not working as long as I set up the current limit afterwards:
dcPowerSession.Outputs[ChannelName].Source.Voltage.CurrentLimit =0.05
thanks,
Shawn