03-07-2009 01:24 PM
03-07-2009 01:45 PM
No. The 6008 can only measure voltage. You can download the user guide and specifications if the one that came with device was lost.
03-07-2009 02:05 PM
03-11-2009 01:56 PM
Hi jmfox180!
Like Dennis said, the only way to do it is using the shunt resistor.
I hope this document can be helpful to you:
http://digital.ni.com/public.nsf/allkb/82508CD693197EA68625629700677B70?OpenDocument
Have a great day!
Oswald Branford
03-11-2009 02:55 PM
jmfox180,
There are several companies that make current transducers. These devices measure the current using a CT or a Hall Effect device and put out a voltage that is proportional to the current. Read output voltage of transducer and scale accordingly. No need to introduce any additional resistance in your circuit.