05-04-2016 10:29 AM
I'm using a usb-6281 daq and keep getting error 89130 when I try to use the counter inputs. The analog inputs and other digitral inputs work fine. I'm testing it using Ni-Max. I assume that if the analog inputs are working then it has nothing to do with resetting the device (which I've tried with no luck). Any ideas?
05-05-2016 12:53 PM - edited 05-05-2016 12:54 PM
So you restart pc, go to max, reset device, start test panel, counter I/O tab, edge counting from any PFI, it gives error, right?
Try different USB, force uninstall/reinstall driver in windows task manager, different PC if possible, MAX Tools -> reset config data. If not, I would say it is bad, contact NI for repairs.
They will also need daqmx version (may ask to upgrade), and windows version.
05-05-2016 01:13 PM
Tried restarting the PC (strictly a wall powered laptop) etc.. No luck. Took the DAQ to another PC (laptop) and it worked fine. Took another DAQ (a 6218), ran the same tasks from Ni-MAX after resassigning them to the new DAQ and didn't encounter the 89130 problem. In other news, it appears the bug that I started with was a result of ground loops in the device I was given to measure. I rewired the device and so far haven't encountered the intermittent signal.