Hi,
Looking at the code you posted it seems you are taking the average of a set number of points acquired and when you do that you are getting a range of values of about 40 microstrain. The Bias current should not be a problem so that would not change the way your measurement behaves. The drift you are seeing over the course of that time could definitely be related to your strain gauge being heated by the continual excitation it is being exposed to. This could result in a slight changes in the resistive behavior of the active elements. Do you know if the Full Bridge configuration is setup in anyway to test for temperature changes in the strain gage element. Another possiblity would be a large lead distance exposing your single ended signal to large noise issues. Differential signals do not have this problem, but single ended measurements don't behave well when dealing with large lead lengths.
Looking at the following documents might give us a start on where we could be going wrong.
This would be a good place to inspect possible noise issues you could be seeing.
The other document has more to do with possible temperature issues with the long term operation of your strain gauge.
I will keep looking into the issue. A good way to test out your device would be to connect a known voltage to the channel terminals and see if the voltage reads in appropriatley. It would be a good step to eliminating any of the internal circutry as responsible for the drift we are seeing.
Have a great day,
Michael D
Applications Engineering
08-07-2007
11:26 AM
- last edited on
07-16-2025
04:25 PM
by
Content Cleaner
08-07-2007 03:11 PM
I want to the test---connect a known voltage to the channel terminals and see if the voltage reads in appropriatley
Can you give more details how to do it? My understanding is connecting S+ and S- to a known voltage. But how can I get voltage readins.
Also what do you mean "Full Bridge configuration is setup in anyway to test for temperature changes in the strain gage element"?
Please suggest. Thanks