Hi,
The board you are using is only programmable using NI-DAQ Version 6.9.3. The device configuration is done through the Measurement & Automation Explorer (MAX). Under Devices and Interfaces you should see the Device listed. Right click on it and select "Properties...". You will get a popup screen with different Tabs
- General> You can configure the device number and test resources for the device
- Analog Input (AI) > you will be able to configure AI mode (Differential or Single Ended) and the polarity if available.
After the config is done, apply your differential signal to one of the 4 Diff channels that are represented by the couples: 0,4 1,5 2,6 3,7.
This should take care of your problem.
Regards,
AlessioD