Multifunction DAQ
daq mx error 200220
hello sir
 i have pci6052E card connected to scxi 1001 chasis and two moduel
1) 1520 with terminal block 1314
 2) 1102 with termional block 1303.
i have built my own program to acquire 4 to20 mA signal from pressure transducer which wowrked fine , butthen i formated my computer and insatlled everything agin then the pregram doednt work
it gives error 200220 .. please see th pics attached. i have atached a snap shot of my program also.
please give some solution.
lalit
(3,678 Views)
Hello lalit,
Since reformatting your computer, have you used Measurement and Automation Explorer (MAX) to configure your SCXI chassis?  From the error message you posted, it seems that this is the problem.
To configure the carrier, open MAX, expand Devices and Interfaces, right-click on NI-DAQmx Devices, and choose Create New DAQmx Device >> NI-DAQmx SCXI Chassis and select whatever chassis you are using.  Then, you will get a screen that allows you to choose what DAQ device it is cabled to and the communicating module slot.  This would be what ever module slot you have the cable connected to.  Click Save and in the next screen your modules should be auto-detected and you can select their accessories.  Then, to make sure that this configuration will run with your existing program, check that the device names match what they were before.  For example, your SCXI-1102 should be SC1Mod2.
I think if you go through the steps I have described, you program should work again.
Laura
(3,655 Views)
