Surprising that the addition of an RC filter fixed everything. One problem I've had with the PXI-6229 is the relatively low source impedance -- only 2k ohms. Was expecting something more in line with a voltmeter, like greater than 10k ohms. So I've had to add buffers to high-impedance sources to compensate. Adding a capacitor could make the signals you are measuring look like a lower impendance (briefly anyway), and so might compensate for the low measurement impedance.
Don't know what the "glitch" problem was from the application note, but maybe charge output from the microprocessor's multiplexer as it switches among channels?
01-08-2007 01:37 PM
01-08-2007 02:02 PM
Thanks for commenting on my misinterpretation of the 6229 input impedance. Saw a reference in the DAQ M Series User Manual mentioning really high input impedance, but also a suggestion that the signal source impedance should be less than 1k ohm for good response in sampling.
I disagree, however, on your comment that an input filter will degrade the subsequent sampling stage. Assuming the sampling multiplexer “breaks-before-make”, having a capacitor on the input signal should only improve the sampling accuracy of a high impedance (but of course low frequency) source. It should swamp whatever capacitance is in the sample & hold circuit of the 6229 or parasitic capacitance. With a higher-frequency, but also high-impedance, source signal probably the only alternative is to insert a follower amplifier as the DAQ User Manual also mentions.