When connecting thermocouples to a DAQ card, my recommendation is to use referenced single-ended mode (as opposed to differential) [1]. This means that each thermocouple only requires a single analog input channel (whereas differential mode requires two channels per thermocouple). The physical connections you would make are: positive lead of thermocouple to an analog input channel and negative lead to the AIGND line.
Your consultant is also suggesting that you use referenced single ended. Differential mode will work, but it requires external resistors to bleed off bias currents in the system that are generated by unreferenced thermocouple.
In SingalExpress, be sure that you're properly setting the channel's terminal configuration mode. Since SingalExpress defaults to differential mode, this is most likely the source of your error.