What noise level should be expected for a K-type thermocouple used with a PCI -6230 DAQ and a CB-37F-HVD breakout board? I am reading the data in with LabVIEW and the Daq Assist module on the thermocouple setting and seeing noise of +/- 1°C. For a K-type thermocouple 41uV/°C is 3 times the expected noise (13 uVrms) of the 6230 if it is running in +/- 0.2 Full Scale. I feel that there should be more stable measurements or is this normal behavior?
http://sine.ni.com/ds/app/doc/p/id/ds-31/lang/en
Thanks,
Dan
Solved! Go to Solution.
05-09-2013 10:43 AM
This level of noise is expected. To improve, you could use a card with a noise filter. A dedicated thermocouple module would also work but this would include a noise filter as well.