03-19-2013 06:15 AM - edited 03-19-2013 06:20 AM
(Broken english approaching)
Hi,
we have to *identical* test stands, each has one USB-9481 for digit output (3 lines of port 0).
Test stand A works without flaws.
Test stand B makes trouble:
Every now and then the error shown in picture 1 occurs (DAQmxWriteDigitalU32).
If you click okay, the MAX's test panels will be opened (picture 2)...!?
The DO-task's timing is "sample on demand", and the samplerate is 10 hz.
Both rigs run WinXP+SP3 with updates and the USB hotfix (WindowsXP-KB969238-x86).
The NI USB devices are connected directly to the port of the computer, no USB-hub.
We connected the USB cables to other USB ports.
We cannot reproduce the error. THe test stand works one day without an error, the other day the errors occurs twice or more.
Both test stands are some hundred kilometers away from each other, so we cannot exhange the USB devices to see, if the error's location changes, too.
Any ideas? I have none. 😕
03-20-2013 04:10 AM
Hi Holwo,
i will look into your request. Could you give me some more information about the system.
Are the Systems Windows XP 32Bit`?
What driver versions are installed? NI-DAQmx??
You can find that info in the "Measurement and Automation Explorer"
Which LabVIEW Version do you use? With our without Service Pack?
Thanks a lot,
Andreas K
National Instruments Germany
03-20-2013 04:24 AM
Hi,
thanks for reply, i've called the NI tech-supp: I'll get 9421/9481 loaners to see if the error is caused by malfunctioning hardware.
Requested info:
03-20-2013 04:51 AM
Hi,
Thanks al ot for the infos so far.
Ah ok great. So you can test it with the loaner.
But what NI-DAQmx Driver Version is used?
Could you tell me that? Maybe an upgrade to a higher
version could fix the problem. You can find the NI-DAQmx
Version in Measurement and Automation Explorer in the section "Software"
and then scroll down to NI-DAQmx ??
thanks a lot
Andreas K
NI Germany
06-05-2013 05:05 AM - edited 06-05-2013 05:08 AM
It's me again (Holwo=hw70).
The customer used the loaners for a while, but there was no change.
Hardware for one test stand (there are two identical test stands):
2x USB-9162 carriers
1x USB-9421
1x USB-9481
DAQmx was updated to version 9.6.2.
Picture "daqmxerr00.jpg" shows the error that randomly occurs -- rarely once per hour, more often it occurs... more often per hour.
That was the only type of error before the DAQmx update.
Now, with the updated version of DAQmx, very, very rare the other errors occur ( --> daqmxerr01/2.jpg).
It's always the digital output tasks that were assigned to the 9481.
We've tested the following:
- Put the 9421 into the carrier of the 9481, and the 9481 in the carrier of the 9421.
- Exchanged the customers' 9481 module with the loaner 9481 -- customers' 9162 carrier.
- Exchanged the customers' 9162 carrier with the loaner 9162 -- customers' 9481 module.
- Exchanged the customers' 9162 carrier with the loaner 9162 -- loaner 9481 module.
- Exchanged the USB cables of the carriers.
- Exchanged the USB ports (front/rear ports of mainboard, OEM Gigabyte)
There are only digital tasks, reading from 2 lines and writing to 2 other lines with 5 Hz, one sample on demand.
I think, the errors show in pictures "daqmxerr01", and "daqmxerr02" are sideffects only.
The main problem is (picture "daqmxerr00") that the hardware (9481) is disconnected from the system. Right after the error occured, there's a tooltip appearing in the right bottom area above the system tray. The tooltip says, that the hardware was installed and is ready for use (sry, no screenshot available).
You can now restart the application (DASYLab), start the experiment/measurement, and every works fine - until that error occurs again.
06-06-2013 02:56 AM
Hi hw70,
in other service requests deleting and recreating the DAQmx Task solved the problem. Maybe you could try that.
Which DasyLab Version do you use? How long are the USB Cables? Is the USB Port a 2.0 Port? If yes is it also
set in the BIOS Settings that it works in the full mode? Which version of "NI System Configuration" is installed?
Can you try the hardware and software on anohter machine so that he can be sure that the USB Bus is working properly.?
Or Maybe the NI DAQmx 9.6.2 is to new for your DasyLab Software?
The driver that is mentioned on the DasyLab Homepage is DAQmx 9.6.1
DasyLab:
http://www.dasylab.com/content/content.php?action=opendoc&sid=555&kat=quick&rcid=27
Driver:
http://joule.ni.com/nidu/cds/view/p/id/3622/lang/de
Or maybe you could get a loaner for the new 1 Slot cDAQ Chassis?
http://sine.ni.com/nips/cds/view/p/lang/de/nid/209817
thanks a lot
Andreas K