I have a DAQ card NI 9203, and I need to use it to collect data from 4 different instruments. Two of the instruments have a maximum sampling rate of 1 sample/ s and two of them have a maximum sampling rate of 100 samples/ s. Is it possible to use the same card?
0 Kudos
Message 1 of 4
(3,512 Views)
Why not sample at the higher frequency ( here 100 Hz) on all four channels??
You can always decimate the slower responding channels, offline
0 Kudos
Message 2 of 4
(3,510 Views)
Thank you for your reply. By decimate, do you mean apply filtering?
0 Kudos
Message 3 of 4
(3,508 Views)
Decimate (in this case) means discard 99 of 100 samples.