Multifunction DAQ
SCXI
Q1: How to get right result at high sample rate with SCXI-1102?
My signal acquire system include:
PXI-1011 mainframe;
PXI-8176 (PIII 1.2G 256MB sdram ) controller;
PXI-6052E 333Ksa/s daq card;
SCXI-1102;
SCXI-1303;
K-type thermocouple;
LabView 6.1;
NI-DAQ 6.9.3..
I want to sample 25 channels together. When sample is set rate as 2Ksa/s, system returns right result. But when I raise sample rate such as 5Ksa/s, system return incorrect values.
I use another system built with DAQCard-6024E, SCXI-1001 SCXI 12 slot chassis and the same SCXI-1102 module. The system works correctly.
How can I get correct result?
Q2:
I want to insert SCXI-1102 programmatically in the program. I often need to change SCXI modules. But now only in the NI MAX could do
it. Is it possible to do this programmatically?
(2,848 Views)
Q1
Are you using bias resistors? You can run into problems when using a floating source in differential mode without using bias resistors. The SCXI-1303 actually has resistor packs that allow you to easily add bias resistors. I would make sure you have those resistor packs installed. The user manual linked below goes over installing the resistor packs.
SCXI-1303 User Manual
http://digital.ni.com/manuals.nsf/webAdvsearch/9DA3B7D5C7F572F986256694004DEFE6?OpenDocument
Q2
Unfortunately, there isn't a way to programmatically add SCXI modules. There is an "auto-detect" feature within MAX that allows for an automatic detection of SCXI modules. You will have to be in MAX in order to add modules.
I hope this helps.
Todd D.
(2,848 Views)
