Multifunction DAQ
SCXI-1102 multiple TC problem
Infrequent user here (unfortunately). Our lab has a setup like this:
SCXI-1303 -> SCXI-1102 -> SCXI-1000 -> NI-6251 -> laptop.
We've been running a student experiment for some time that uses a single thermocouple wired to channel a0, and I'm trying to expand it to multiple TC 's. I wired a second TC to channel a1, created a suitable task in NIMAX,  and started it running in continuous sample mode.
The result was a two-trace chart as expected, but both traces indicated the temperature on the first TC. Applying heat to TC0 caused both traces to go up; applying heat to TC1 had no effect.
However, if I create a task using either one of the TC's, its single trace responds normally to heat application.
I tried several other combinations of channel connections, and the behavior is the same: the temp reading of one channel dominates the other in two-TC mode.
Suggestions? TIA...
rj
Message 1 of 14
(5,512 Views)
rj
What are those?
Message 2 of 14
(5,490 Views)
SCXI-1303 terminal block
SCXI-1102 thermocouple module
SCXI-1000 chassis
NI-6251 mass term
Is this not the right subforum?
rj
Message 3 of 14
(5,485 Views)
It sounds like a Task may not be configured properly in NI-MAX.
Message 4 of 14
(5,471 Views)
OK, here's a copy of the task. This one is set up to use channels ai2 and ai4, and the channel that dominates the chart in this case is ai2.
Message 5 of 14
(5,465 Views)
You wrote in your first post that you were using channels ai0 and ai1.
Where did ai2 and ai4 come into the picture?
Anyway, you have some other things and devices going on in that config file I don't quite understand.
I would suggest simplifying to just the thermocouples.
In addition, if you have a USB 6251, I don't understand the SCXI stuff.
Message 6 of 14
(5,457 Views)
I tried several other combinations of channels to see if they made a difference, but they didn't. Channels ai2 and ai4 act just the way ai0 and ai1 did.
The NI-6251 is simply interfacing the SCXI chassis to the laptop, replacing the PCMCIA card we used to use when our laptop was an XP machine.
What are the "other things" you mention? The chassis also has a strain gage module in it, but nothing other than the thermocouples is connected to the SCXI-1303.
Message 7 of 14
(5,443 Views)
You should be able to connect to the USB-6251 directly. I don't understand the purpose of the SCXI equipment.
The USB-6251 has 16 analog inputs https://www.ni.com/en-us/support/model.usb-6251.html
Message 8 of 14
(5,434 Views)
Do not use NI 6251. Instead connect output of 1600 module direct to usb port of laptop. Use Ain0 and Ain1.
Regards
Jamal
Message 9 of 14
(5,350 Views)
Regards
Jamal
Where does he write that he has a "1600 module" ?
Message 10 of 14
(5,346 Views)
