Hi,
The range inside of MAX should read +/- 10 V, the range of the card with a gain of one.
To see a larger range in your application, you must configure the gain to be less than one. For example, the NI 6110 supports gains of 0.5 and 0.2, which allow ranges of +/- 20 V and +/- 50 V. (Remember that the card supports a maximum of 42 V.)
Depending on the range of your input signal, you would want to use the following gain setting:
| Range | Gain Setting |
| -50 to 50 V (not to exceed +/- 42 V maximum) | .2 |
| -20 to 20 V | .5 |
| -10 to 10 V | 1 |
| -5 to 5 V | 2 |
| -2 to 2 V | 5 |
| -1 to 1 V | 10 |
| -500 to 500 mV | 20 |
| -200 to 200 mV | 50 |
To set the gain in LabVIEW, use the Hardware Configure VI and set the input range of the signal you want to measure, and the VI will select the best gain. Refer to the link below for an example.
To set the gain using NI-DAQ Function Calls, pass in the gain as an integer. If the gain is 0.5 or 0.2, pass in a -1 or -2 respectively.
Hope that helps
Sacha Emery