Multifunction DAQ
NI9232 input test using resistance.
I am trying to measure the acceleration sensor value of PIEZOTRONICS by mounting NI9232 to the C-RIO module. I want to see the performance of this module without an acceleration sensor, but I want to see the voltage that this module receives without an acceleration sensor. Is there a way? The way I thought about it is to measure the voltage by connecting about 2.5KΩ directly. Can I do this?
(1,762 Views)
You can, as long as you have a signal generator like a function generator. This module measures voltage from a source, a resistor being a passive element cannot be measured using this module.
Santhosh
Soliton Technologies
(1,739 Views)
Should I connect it and measure it?
(1,712 Views)
Yes, that is correct.
Make sure you're testing the module with sine signal as vibrations in real-world are composed of sines.
Santhosh
(1,708 Views)
For a test you can test the IEPE input with a 2.5k or 2.7k resistor. (assuming 4mA supply current)
To measure the transfer function, two methods (IEPE simulators) are presented here:
https://www.ptb.de/cms/fileadmin/internet/dienstleistungen/dkd/archiv/Publikationen/Richtlinien/DKD-...
To measure noise performance, just keep in mind, that the actual IEPE circuit can damp the current source noise..
so the easy way for input noise tests and tranfer function: turn off the current source, but keep the high pass filter.
for current source noise the 2.5 k is the way to go.
Greetings from Germany
Henrik
LV since v3.1
(1,698 Views)
