Multifunction DAQ
NI-USB 6008 reading in voltage problem
Hi
I am trying to program a USB-6008 on a mac. When I connect the Analog input to the 5V output I get a reading of 3.67volts, while on an osscilascope I read a voltage of 5volts. Is this to be expected? Do I need load resistors? Also I get the same effect when outputting voltages for the analog output.
(3,968 Views)
I have also tried this with 2.5V and get 1.1 volts
(3,960 Views)
Accepted by topic author zepp2
Hi Zepp2,
Thanks for your post. Which pins are you connecting together on your device? Dependant on the type of connection you have the wiring may not be quite right.
I have found a few online documents which hopefully should help you out:
Field Wiring for analog signals
USB-6008/6009 Getting Incorrect Voltage Readings
USB-6008/6009 Analog Input Reads Approximately -1.4V Offset
Please let me know if any of these help you out.
Ben B.
National Instruments UK & Ireland
"I've looked into the reset button, the science is impossible!"
(3,932 Views)
