NI-9237 "external" excitation

Hello to all,

I need to acquire signals from a load cell (full-bridge) and 3 displacement transducers (full-bridge too) using only a NI-9237.

The problem is that the load cell bridge is already feeded by another module at 2.5 Volts (unfortunately has to be so).

I wonder if it's possible to NOT connect EX+ ed EX- and simply read the AI+ AI- voltage on that channel.

 

 

Thanks,

Manga

 

 

0 Kudos
Message 1 of 4
(4,773 Views)

hi manga,

 

I post you this information that could be helpful:

 

 

 

"When external excitation is enabled on the NI-9237, the excitation ± terminals (pins 6 and 7) disconnect from the internal cDAQ backplane source and connect to the 4 terminal EX ± AI Bridge on the front of the device.  With external excitation enabled, a 2-10 volt external power supply needs to be connected in order to provide excitation. 

If the external power supply voltage is less than 2 volts, you will see inconsistent voltages measured across pins 6 and 7 (EX ±).  For example, with 0.5 volt excitation applied, the measured voltage might be 4.5 volts.  There is no workaround for this; the minimum voltage the 9237 accepts is 2 volts.  This is a constraint placed on us by the ADC.

With internal excitation, there's a 150mW maximum power draw limitation.  With external excitation, we didn't calculate or test for the maximum power draw as we couldn't think of a valid customer use case that would put the module at risk.  As long as the customer is using strain or bridge modules, we should never draw too much power through the module. "

0 Kudos
Message 2 of 4
(4,737 Views)

So, I should be able to make measurements on the AI0+/- terminals, with nothing wired to the Ex0+/- terminals, as long as I set the excitation mode to 'external' and apply >2V to Vex+/- terminals, to power the ADC?

0 Kudos
Message 3 of 4
(4,377 Views)

Confirmed!

 

Using full bridge load cell. (3mV/V signal; 5000lb capacity)

Tested with excitation ranging from DC5V to DC 14.5V

Created task in Max; Signal type: 'Custom Voltage with Excitation

Chart shows: 10lb test sample; then 54lb test sample

 

 

9237 with external excitation hw.jpg

9237 with external excitation sw.jpg

Message 4 of 4
(4,349 Views)