01-05-2017 05:07 AM - edited 01-05-2017 05:08 AM
I am measuring temperature with thermistor sensor using NI 9219 AI module on a cDAQ 9178 chassis. There is no issue when measurement is done using 2-wire mode and displayed result is in complete agreement with calibration report provided by Thermistor manufacturer. However we want to take measurement over a long distance. To get accurate result, we used 4-wire resistance measurement scheme. But the displayed result shows Zero resitance (short). The connections are proper and checked multiple times. I am not able to figure out the error. Whether NI 9219 cannot measure Thermistor resistance in 4-wire mode.
At 25 deg C, Rthermistor = 10k.
Need help.
Solved! Go to Solution.
01-05-2017 09:32 AM
Hi, what do you mean by "long distance"? What type of wires do you use?
Does the connection look like this?
Also - the program you sent is configured to use 2-wire, not 4-wire method.
01-08-2017 11:45 PM - edited 01-08-2017 11:46 PM
My bad.
I switched the Pin 4 connection with Pin 6. I will try after rectifying it.
Thanks for your nice connection visual.
09-25-2018 06:01 PM
@arihant
were you able to get this working? What is the accuracy you are getting?
Thanks
Bhavin
04-12-2020 01:35 AM
Sorry as I could not reply.
Yes, I was able yo do it.
04-14-2020 10:09 AM