Multifunction DAQ
Measure Frequency gives wrong value
I recently got the USB-6210 and I also have the pci-6251.
I'm trying to measure the frequency of a TTL signal. I'm using a signal generator as my signal source.
I'm using a continuous frequency measurement task and averaging the results. When I look at the signal on an oscilloscope and the pci-6251 I get the correct frequency but when I use the USB-6210 I keep getting a higher frequency.
For example:
50kHz --> 50.0313 kHz
100kHz --> 100.125 kHz
500kHz --> 503.145 kHz
Can anyone explain what is going on? Is there a way to fix it?
Eyal
(3,254 Views)
I forgot to mention that I don't think that the +-1 clock cycle error explains the results since I'm averaging N samples (N is very big ~ 1Mcounts) and it should cancel out. Although it is suspicious that the results I'm getting correspond to exactly -1 clock cycles per measurement (I'm using the 80MHz clock).
(3,253 Views)
Hello Eyal S,
I have a couple of suggestions for further troubleshooting this issue.
  1. Try a frequency measurement of the internal 100 kHz timebase.
  2. Generate a pulse train from the USB-6210 and read the frequency using the PCI-6251.
  3. Verify that the behavior is consistent between the 2 counters on the USB-6210.
These tests will help to shed light as to what is going on with your USB-6210.
Dan King
(3,229 Views)
Hi Dan_K,
Thanks for your reply.
I actually added one cycle to all of the data points and the problem is fixed. There must be a bug somewhere...
Eyal
(3,220 Views)
