02-06-2014 06:21 PM
I have a signal that is 60mVpp AC. I want to measure the Vrms in LabView. What hardware could I use for this if I am trying to detect a change as small as 0.1mVpp.
02-06-2014 09:46 PM
02-07-2014 01:21 PM
Dennis is right, but take a look at the NI PCIe-6320. It should fit your application.
02-28-2014 12:50 AM
This additional information will help us narrow down the options:
-sample rate
-do you need accuracy or precision? Sounds like precision ('detect a change'), but just making sure
NI's standard E/M/X series 'DAQ' cards likely won't get you 100uV accuracy (not to be confused with resolution).
My first recommendation is a DSA card, such as the PXI-4461, or the NI-4071 DMM. I can make a much more informed decision with more information.
02-28-2014 01:59 AM
@Now_With_Underscores wrote:
This additional information will help us narrow down the options:
-sample rate
-do you need accuracy or precision? Sounds like precision ('detect a change'), but just making sure
NI's standard E/M/X series 'DAQ' cards likely won't get you 100uV accuracy (not to be confused with resolution).
My first recommendation is a DSA card, such as the PXI-4461, or the NI-4071 DMM. I can make a much more informed decision with more information.
Sorry,
here is the spec of the 6251 (M?) DAQ
I would expect that you can measure a 60mVrms AC signal in the +-100mV range and do better than 1100ppm absolute accuracy 🙂
But if the poster can give more information on frequency range , surrounding conditions (Lab, industrial , outdoor,..) , traceability, number of channels etc. there would make a difference in spending 600 or 10k $Unit 😉
BTW: 60mV AC in audio range with 0.1% relative change can be done with the line audio input of the soundcard 😄
02-28-2014 10:49 AM
Hey Henrik, yes, the 6251 is an M series DAQ device.
Let's talk specs ;)...
If you read all the fine print, you'll need to average 100 samples to achieve the specified uncertainty. You'll also need to stay very close to the external and internal self-cal temperatures, as those add uncertainty very quickly as temperature changes. Assuming you stay within ±5°C of external and internal self cal, AND you average 100 samples, then yes, you can achieve 83.7uV* uncertainty in the 100mV range with a 60mV reading.
However, if our mutual comrade is looking for 100uV uncertainty, then I propose that 83.7uV uncertainty is not enough; typically, we want 4x less uncertainty on the measuring equipment.
If our mutual comrade is most interested in precision, then I propose looking at any of the various DSA cards (4431, 498x, 446x, etc). If we look at the 4461, we can see that our gain error is much smaller.
Overall, I agree we need more information from our customer.
Have a great day
*Using equation in spec sheet:
60E-3*(150E.6+13E-6*5+1E-6*5)+100E-3*(150E-6+90E-6)+15E-6*3/sqrt(100)=83.7uV