02-17-2013 12:06 PM
When performing self calibration with an SCXI chassis configured, error -201295 is returned.
02-18-2013 07:03 AM
It's clear that leaving the signal cables connected during self-calibration is causing this error. According to the DAQmx Help, Device Calibration Signal Connections for X Series Devices, the first recommended step is to disconnect any external connections or circuitry to your device.
The help also states that the analog output values change during the calibration process and according to this discussion forum, others have seen this error when the analog output channel is being interfered with in some way. If you have connected signals to the AO, it's likely interfering with the self-calibration process and disconnecting them and the chassis will lead to a normal self-calibration.
02-19-2013 08:25 AM
The steps in which you refer apply to external calibration and not self calibration (read the second paragraph carefully). The M and X series devices were designed such that they can be calibrated in situ. Otherwise, including programmatic self calibration methods is pointless.
Thanks for the tip on the AO channels. In our case, they are only connected to high impedence loads, so there shouldn't be interference issues, but at least I know the nature of the problem now.
02-19-2013 05:48 PM
Have you tried running this on another computer? The fact that your values still update after the error is thrown leads me to believe that this is a driver issue. Try doing a Database reset in Measurement & Automation Explorer. You can do this by going to Tools>>Reset Configuration Data.
Regards,
-Travis E
02-21-2013 11:21 AM
This is currently happing on 3 seperate installations, meaning 3 cards installed 3 different servers. I will try the database update and report back...
02-22-2013 01:00 AM - edited 02-22-2013 01:03 AM
According to my experience, and like mentioned by CindyShark, the self-calibration often generates an error if the cable(s) is(are) not removed. This depends on the signals.
Therefore, I have already thought to post an idea in this thread to ask NI to change this behavior of their cards (next generation)... but haven't done it yet.