Multifunction DAQ
Large offset values for NI 9222
The initial readings on select channels of my NI9222 and NI9220 DAQ modules are significantly offset from the expected 0V value that I observe (within a reasonable tolerance) on the remaining channels.  My DAQ set-up is composed of 3 modules (NI 9403, NI9222, NI 9220) that are connected via a NI 9188 chassis.  I am testing my set-up with all the modules connected and without hooking up any of the respective AI+/- pins.  I attached a screenshot of a test that I ran on the NI 9222.  Below is a list of all the initial readings for my DAQ modules.  I listed the channels with the large offset voltage that are causing me concern in bold.
NI 9222:
ai0-  10.6 V
ai1-  10.6 V
ai2-  1.97 V
ai3-  0.29 V
NI 9220:
ai0:    -0.002    V
ai1:    -0.003    V
ai2:    -0.0005  V
ai3:    -0.002    V
ai4:    -0.009    V
ai5:     0.0004  V
ai6:    -0.0005  V
ai7:    -0.001    V
ai8:    -0.003    V
ai9:    -0.001    V
ai10:  -0.004    V
ai11:  -0.004    V
ai12:  -0.009    V
ai13:  -10.47    V
ai14:  -0.002    V
ai15:  -0.010    V
Any information on why this might be happening or how I can resolve this issue would be much appreciated.
Max
(4,352 Views)
Hi wmm26,
Have you tried writing a value to these values? Use a multimeter to distinguish the value going in versus the value you are observing in MAX.
Also have you tried using a mulitmeter to measure the input values?
Also have you tried resetting these devices as well?
Have you tried swapping the slots which each module is in?
Paolo F.
(4,310 Views)
Hey Paolo,
I was able to resolve this issue.  The step that helped me was using an oscilloscope to compare measurements that I observed using the MAX.  I assumed that the ground from the power supply would be used as a default internal ground.  Connecting the AIO- terminal to the grounded end of my oscillascope was a sufficient reference to re-calibrate my channel readings at 0 volts.  I found it odd and slightly alarming that only certain channels exhibited this feature and at different levels.  Why is this the case?
Max
(4,306 Views)
Hi Max,
It seems that the 9222's channels are floating relative to earth ground as they channels are isolated from each other.
I found it in the specification sheet found here:
http://www.ni.com/pdf/manuals/374223b.pdf
We recomend you ground these channels with the ground used for the devices in the "-" terminal as the measurements that are expected is differential.
Paolo F.
(4,263 Views)