Multifunction DAQ
Input range binary
Hi, I am collecting data using a M-series 6221.  Because I am collecting a lot of data, I am collecting and storing it in binary format.  The card is 16 bit, and I have the voltage range set up to +/- 10 V.
So, when converting the binary data back to voltage, I need to multiply my binary number by 20 V  / 2^ 16 bits to get the binary number back in bits.  But there is a problem here....when I apply 5 V and measure it, I only read ~ 15 000 bits instead of 16 384 (converting to ~4.7 V).  I have tried this with multiple cards.  The number seems to be off by ~ 8-9% each time at different voltages.  I am runnning in differential mode.  Am I doing something wrong?   Is the range of the bits really larget than +/- 10 V?
Michael
(3,635 Views)
Post the part of your code that is doing the conversion to bits.
(3,619 Views)
Hello Michael,
The working range for the 6221 (6221 Detailed Specifications) series is +/- 11V.  This is to allow a little bit of calibration room and guarantee that you can measure the full range of +/- 10V.  If you are using unscaled or raw data, then you need to use +/- 11V range.  If you use this value in your calculations, then you should be able to convert the data correctly.
Respectfully,
Rob F
Test Engineer
Condition Measurements
(3,614 Views)
You can't just assume ±11V. The actual scaling coefficients for the board are stored in its EEPROM. You can access them through a property node. I can't remember exactly how to find it at the moment.
Chris
(3,601 Views)
