Multifunction DAQ
Inaccurate peak to peak voltage
I am using an usb 6358 DAQ.  In the process of troubleshooting something in the field we were applying known signals to the AI of the DAQ.  For DC signals the DAQ reported accurate values (both through the data collection VI and through MAX).  However when we applied a sine wave we got twice the peak to peak voltage.  For example if we applied a 1 Vpp signal (which should be -0.5 V to 0.5) we would see 2 Vpp via the DAQ (again both through the data collection VI and through MAX).  We tried a range of frequencies (1 Hz to several kHz) and got the same result.  I am sure there is a simple explanation but it escapes me right now.
Thanks.
Message 1 of 12
(5,065 Views)
Just some quick thoughts without knowing yours system.
Is there a scale being applied?
Is the peak-to-peak measurement incorrect (assuming coding was involved to calculate it)?
Is there a buffer that is applying a gain in the circuit?
Message 2 of 12
(5,062 Views)
I am viewing the signal so I am not calculating the peak to peak, just reading it off of a graph.  I have no gain settings for my DAQ.  I have tried changing the range (which affects the internal gain and resolution) but this does not affect the reading.  I am using a differential measurement (no other choice for this particular DAQ) and plugging directly into the function generator.
Another piece of the puzzle:  The function generator (a modern digital model) reports that it is outputing a 1 Vpp signal and shows a nice little graph with a sine wave oscillating b/t -0.5 V and 0.5 V (as you would expect).  We checked this on an o-scope and it was indeed within this range.  When I read the Vrms directly from the o-scope using a multimeter, I get a Vrms of 0.7071 which matches a Vpk of 1 V which would then imply the signal really is 2 Vpp.
Message 3 of 12
(5,060 Views)
Is the AI GND connected to the system ground?  Even though you are using differential measurements, you still need to have grounds common.  I've seen floating grounds do some really weird things.
Message 4 of 12
(5,056 Views)
I am putting the signal into AI+ and the function generator ground to AI-.
Message 5 of 12
(5,047 Views)
What is this source? What does it expect to be terminated with? Sounds like an impedance mismatch where the source might be expecting 50 ohms.
Message 6 of 12
(5,045 Views)
I've been wondering about that.  I connected the function generator directly to the o-scope via a 50-ohm BNC cable and set the o-scope channel to 50-ohms.  The output on the function generator is labeled as 50 ohms.
Message 7 of 12
(5,042 Views)
There you go, then. The DAQ card and DMM are high impedance. The setting of the source is double when you don't terminate in 50 ohms. If you were to change the scope and use a regular probe in high impedance mode, you would see exactly the same thing as the DAQ card.
Message 8 of 12
(5,039 Views)
yep, slap a 50Ω across the terminals.  I've run into that problem more times than I care to admit.
Message 9 of 12
(5,036 Views)
Thanks for the feedback.  I don't quite understand why the low impedance would result in double the voltage.  We're in the middle of tesing now but I will give this solution a try.  Thanks to everyone for their input.  I'll give you guys a tenative "Kudos" until I can verify!
Message 10 of 12
(5,031 Views)