Multifunction DAQ
I cannot read thermocouples on connector 1 of the PXI-6225 M-series board.
I am trying to read 30 thermocouple measurements in differential mode.  I am using AI0 as the CJC.  Attached are the connections I am using.  The negative inputs are all connected directly to AI GND.
I have no problem reading  the thermocouples in Connector 0 (AI1 through AI 7).  I have connector 1 setup in the direct feed through mode (switch 1&2 right and switch 3 up).  In MAX I get a normal reading but it does not vary by heating the thermocouple.  In my actual program it is as though there is nothing connected. Any ideas as to what I may be doing wrong?
Mark
(3,615 Views)
Short cut?  If you short cut a Thermocouple input you will read the CJC temperature, if you card could not detect a short cut.
Greetings from Germany
Henrik
LV since v3.1
(3,588 Views)
Accepted by topic author Mark88
Mark,
Thanks for posting. I believe that the issue you might be seeing is related to how the SCB-68 must be configured in direct feedthrough mode to be used with connector 1 on the PXI-6225. When the switches on the block are set for direct feedthrough, the board is not connected to the CJC. Not having a connection to a CJC on that connector could be causing this behavior.
One potential workaround you could try is to connect a temperature sensor to AI16 and follow the steps in this KB, using that channel instead of the CJC channel. Let me know if this helps!
Joe S.
(3,576 Views)
I do not know what the problem was for sure.  But I think the PXI system I am using needed to be reset after I changed it to direct feed through.  I did this previously but with the power still on.  This morning I shut it off and unpluged it from the wall and now everything seems to work fine.
Thanks for your help
(3,571 Views)