Ordinarily, when you want to generate a DC voltage you just use an On Demand (no hardware timing) task. That means you don't use the Timing VI at all. However, since you want to use the ao/StartTrigger to trigger the AI task, you must use hardware timing so that there will be a ao start trigger pulse. Since you must call the Timing VI you must set a sample rate. Unless you are using the ao/SampleClock to time something else, it doesn't matter what sample rate you choose. Just use the default of 1000. Unless you are using the ao/SampleClock to time your ai scan, it will have no effect on your ai scan. You can leave the sample mode on the ao timing vi to finite. But be aware that the samples per channel input on the ao timing vi will determine how many times the data in your 2 point buffer is generated. The default is 1000. This would write out the same value to the ao channel 1000 times. Again, this is probably harmless. But you could just as well set this input to 2, which is the minimum.