Hi Everyone this Baris From Turkey
I have NI4497 daq card for reading Voltage from sensors. I am using MAX to configure it. One of the Channel Aio is connected to the Sensor and when I measure it with a voltmeter I am reading 3.4V then when I try to measure same sensor for using MAX software in the test Panel I am getting same voltage but on the DAQ task for import it to the Signalexpress... While I run the task the value staring from 3.4 v going the as you can see on the picture.
Why Voltage Value going Down?
For Voltage must use DC what should I do?
Please Help me..
09-26-2012 11:02 AM
Did you configure the channel to be DC coupled? It looks AC coupled to me. The specification claims that the 4497 has channel selectable coupling.
09-26-2012 11:56 AM
Yeah i think it looks like AC coupling when i Open the Test Panel i return the Coupling AC to DC but is there any way to change it DC Coupling
09-26-2012 12:13 PM
Hi Baris,
What version of NI-DAQmx are you using? In NI-DAQmx 9.5 and later the DAQ Assistant allows selecting DC coupling for the NI 4497.
Brad