Hello Stefan,
The M Series User Manual provides the following description of the D GND signal:
"Digital Ground—D GND supplies the reference for
P0.<0..31>, PFI <0..15>/P1/P2, and +5 V. All three
ground references—AI GND, AO GND, and
D GND—are connected on the device."
In addition to these internal connections, the ground signals are also connected to the power supply ground (earth ground). In most cases, depending on your system, the chassis will also be connected to earth ground. So, though the ground signals are not directly connected, they are connected through earth ground. Sharing this common ground is important to reduce ground loops in your system. As discussed in this document, ground loops can induce noise and can cause significant drift on the signal. This topic is also discussed in another discussion forum here. Could you provide some more information about your system? Are you using PCI devices in a desktop PC? Or are you using PXI devices in a PXI chassis?