Multifunction DAQ
Calibrating K-type thermocouples
I've been trying to use k-type thermocouples to measure thermal cycles in welds, but it seems that the temperature readings are significantly lower than the actual temperature (30-50 degrees C). I'm using an NI 9213 module with the DAQmx block. Is there a way to calibrate the thermocouples, and if so, how? I've attached screen shots of the block diagram.
(861 Views)
Well, the TC will report the temperature he feels, that migth or migth not be the temperature of interest 🙂
30 deg offset could be a false cold junction (CJ) compensation.
If you shortcut the TC input with a short wire , you should read the CJ temperature.
A calibration is just a comparison of your setpoint (temp) and the readout.
Depending on the number of calibration points you have, you can do a compensation of the read value.
like: one point: offset from cal and use the slope of the TC standard
two and more points allow a polynominal correction.
Greetings from Germany
Henrik
LV since v3.1
(818 Views)