Multifunction DAQ
Agilent 34970A Error Message
Hello, I'm using Agilent 34790A with EZ Temperature.vi. I assumed that everything is configured with the instruments but when I click RUN, I get one "beep" and an error message (err -420). Does it means that I still need to configure the vi? Any help would be appreciated. Thanks!
SW & HW
LV8.5.1
Agilent 34970A Data Acquisition Switch
Agilent 34901A Module
GPIB
J type TC
______________
KowdTek
LabVIEW 2009
One Step At A Time, Maybe Two...
(4,850 Views)
I don't remember what error 420 is. Can you provide the explanation from the manual? It might jog my memory. I no longer use it but I did for several years.
p.s. You posted to the wrong board. this is for NI DAQ boards. Next time, post to the instrument control board.
(4,849 Views)
Query
UNTERMINATED
The instrument was addressed to talk (i.e., send data over the interface)
but a command has not been received which sends data to the output
buffer. For example, you may have executed a
CONFigure command
(which does not generate data) and then attempted to read data from
the remote interface.
Sorry for posting in the wrong board.
______________
KowdTek
LabVIEW 2009
(4,845 Views)
I don't remember ever getting that error but the driver has been changed numerous times since I started using it. My recomendation would be to start small and run the functions one at a time. Start with the initialize function and then see where the instrument is generating the errror. I don't have the driver installed here but I know there are other top level examples. You could try running one of those.
(4,842 Views)
Ok thanks for your replies.
______________
KowdTek
LabVIEW 2009
(4,840 Views)
I just tried and I'm getting the error right @ Visa Read and the error changes to err -103
Here's the explanations:
-103
Invalid separator
An invalid separator was found in the command string. You may have
used a comma instead of a colon, semicolon, or blank space – or you may
have used a blank space instead of a comma.
Examples: TRIG:COUNT,1
or
CONF:FREQ 1000 0.1
______________
KowdTek
LabVIEW 2009
(4,838 Views)
That's odd too. I never experienced that error and I used the 34970 in several different variations for a number of years (but not the thermocouple, though). I'll try to look at the manual within a day or so and the latest version of the driver. I'm not sure what you can try in the meantime. One thing that comes to mind is to verify that in MAX, the GPIB controler is set to send only EOI.
(4,836 Views)
