Multifunction DAQ
AI works for a while, then returns fixed value until reset
This is probably a simple newbie thing.
I've got a USB-6215 and am using NIDAQmx Base. With our software, or with the acquireNScans example, I'm getting a certain number of reasonable results, and then at some point I get a fixed value which is outside the configured voltage range. That continues across program runs, and only stops when the 6215 is unplugged and reconnected.
For example, if I run acquireNScans once, it works as expected, collects 1000 samples and prints out the first ten, which are reasonable-looking values. I modified the code to print  out all 1000, and all 1000 look reasonable.
Acquired 1000 samples
data[0] = 0.336264
data[1] = 0.264918
data[2] = 0.148858
data[3] = -0.003368
data[4] = -0.174993
data[5] = -0.342672
data[6] = -0.486021
data[7] = -0.588929
data[8] = -0.638904
data[9] = -0.632329
data[10] = -0.568874
But if I run it again without unplugging/replugging the 6215, I get
dhc023738:ai jon$ ./acquireNScans
data[0] = 10.883438
data[1] = 10.883438
data[2] = 10.883438
data[3] = 10.883438
etc
Note that this is on a -10v-10v range.
I assume there's something I'm not doing, and the example isn't doing, such as clearing a buffer.
 Any tips?
Jon
Message Edited by JonHendry on 09-28-2009 05:24 PM
Systems Manager / Programmer
Maunsell Lab
Dept. of Neurobiology
Harvard Medical School
(3,239 Views)
Good Evening Jon,
What signal do you have connected to the input which you are measuring?  What values should we expect to see from it?  A floating input is likely to eventually rail (go to the high or low supply voltage).  If you do not have anything connected to this input, I would expect you to see the results that you see:
  • The data looks "right" (about zero) for some amount of time.
  • It then ramps up or down (to +10 V or -10 V, respectively).
  • The only way to get it back to around zero is to unplug the card so it loses power and the inputs discharge.
If there is nothing connected, try grounding the input to an Analog Input Ground if you are measuring in Single-Ended mode or connecting the + and - inputs of a differential input to each other.  This should provide a consistent zero reading.
(3,215 Views)
