Hi Paul-
It sounds like you're running into the issue described in this KB. Basically, you need to make sure in your application that you specify the correct terminal configuration; 'default' is the default (strangely enough ) and results in a 1.4V offset if you use RSE connections.
Hopefully this helps-
08-30-2006 08:39 AM
Thanks for the KB it didnt quite solve my problem. I have configured the AI to be a differential AI not the RSE as mentioned in the KB. Also when the signal is not attached there is no voltage difference but when the signal is wired to the 6009 I see a +1.4V offset. I dont see this from a battery which measures the approperate ~1.5V expected value. Very strange and has stumped me. Since the offset is constant I am factoring it out but this is obviously not the ideal situation.
Paul.
08-31-2006 02:25 PM
It is curious that you are able to read the battery but not your test signal. Could you provide more information about the nature of the signal you are trying to measure? Is it floating or ground referenced? Have you tried bias resistors? You can find more information about analog signals and bias resistors here.
Do you see the same offset voltage on each of the channels? Since the offset is 1.4V I believe the issue is still related to the input circuitry. If you connect one of your analog output channels to your analog input can you measure the full range of voltages accurately by running a test panel and updating the analog output voltage from 0-5V? This is easier if you select the test panel from the popup that appears when you first plug it in, and then run a second test panel from Measurement and Automation Explorer. This way you can have one continuously reading the analog input and also be updating the output at the same time.
Jennifer O.
09-01-2006 08:20 AM
I am looking into the input circuitry. I figure it is something to do with this. The input comes from a analog PMT module, and some electronics wired by a fellow scientist at work. I will see if I can get more information. The PMT is measuring the scatter of an aerosol particle as it passes through a laser beam. The PMT should be a slight negative voltage and go more negative as the scatter increases when a particle passes through the beam. The voltage should be from 0 to -5V depending on the gain setting of the PMT's amplifier. I am a CS guy not an EE so I will look at the bias resistor tutorials. Thanks.
Paul
12-06-2006 08:56 AM
12-06-2006 09:08 AM
12-06-2006 12:48 PM
12-06-2006 05:37 PM - edited 12-06-2006 05:37 PM
Sorry for the confusion on the specifications. The USB-6008/9 User Manual and Specifications document is available online through the "Resources" tab under the product page. The Analog Input Circuitry is show on page 11 and shows the 2.5V Ref feeding the resistor divider network used to scale and shift the input voltage. In addition the input impedance is listed as 144kOhm in both the User Manual as well as the data sheet. As far as isolation, the USB-6008/9 does not offer isolation, but another NI USB bus-powered device (the USB-6215) does offer isolation. Also the USB-621x devices have >1GOhm input impedance which is great for high-impedance sources. The full specs for these devices can be found here.
The USB-6008/9 are great for some applications, but if you need the performance (isolation, speed, input impedance, etc.) the USB-621x is a nice step up.