5102 error

When I use the LV sample code to collect data, there is error when try to get the cluster data(niScope Fetch(poly).vi).  Any suggeston?
Also I tried to set the analog triger with certain level, will it affect it?
 
The error message is,
 
Error occurred at:  niScope Fetch.vi:3
<err>Driver Status:  (Hex 0xBFFA2003)
Maximum time exceeded before the operation completed.
Possibly no trigger received. Buffers done: 0, Points done: 0.000000.
 
We are using winxp and LV8.2. Will it be supported by the driver? The tool of ni scope also not works.
Is there any patch of driver?
We tried on win2k and LV7.1, the card works fine.
0 Kudos
Message 1 of 2
(2,731 Views)
If the problem appeared when you tried to adjust the analog trigger it is very likely it's not triggering.

Try setting the trigger to free run (or whatever they call it). It should work. Switch back to analog trigger and insert a probe so you can see whether the trigger condition has been met.