Multifunction DAQ
15V power supply calibration ELVIS circuits trainer
Is there a way to adjust the voltage from the +15V power supply? I'm getting 15.5V and would like to adjust it back down to 15.0.
(3,812 Views)
Are you reading that with no-load?
Throw something at it to load it down a bit then take a measurement.  Regulators usually need some load (more than the meg- and gig-ohm Z that most meters present).
(3,802 Views)
Yes, the power supply has a load. There are numerous series/parallel circuits that the students build and measure voltages and currents.
(3,754 Views)
Accepted by topic author Justin.e.snodgrass
Hello Justin,
If you look at the specifications for the ELVIS, on page 10, you see that we specify the +15 V power supply with an uncertainty of ±5%, so it’s expected for your power supply to go from 14.25 V to 15.75 V.
That being said, as this is a fixed power supply, there is no way to bring the voltage down from 15.5 V to 15 V without using external circuitry. Please reference this thread for further information. I think using an OpAmp will be the best solution for you. For your convenience, I found the circuit you want to make on this datasheet, page 7, which is a Non-Inverting DC Gain amplifier. Because the ELVIS already has a breadboard, the implementation should be really straight forward.
Daniel REDS
RF Systems Engineer
Help us grow.
If a post solves your question, mark it as The Solution.
If a post helps, give Kudos to it.
(3,741 Views)