Querying Agilent E4407B with No Results...Help!
I'm very new to VB .NET and Measurement Studio, so please forgive me if this gets confusing.
I want to create a very simple program using the VISA class library that will query a spectrum analzyer and display the results of that query in a textbox.  The code is as follows:
Private Sub Button1_Click(ByVal sender As System.Object, ByVal e As System.EventArgs) Handles Button1.Click
     Dim mySession As MessageBasedSession = CType(ResourceManager.GetLocalManger().Open(GPIB0::18::INSTR), MessageBasedSession)
     TextBox1.Text = mySession.Query("IDN?\n")
     mySession.Dispose()
End Sub
When I run the program, I get a "Query Unterminated" on the spetrum analzyer and then the program terminates due to a timeout experiration on the Query().  What is going on?
Also, what is the difference between the synchronous/asynchronous methods (i.e. BeginRead/Read, BeginWrite/Write), and which ones should I be using?
Once I can understand the fundamentals, I would like to create another program that will capture the display on the specan and plot the points on a WaveformGraph.  Here is a snippet of code that I will be using:
     Dim s As String
     mySession.Write("TDF P;TRA?")
     Dim traceA As String = mySession.ReadString()
     Dim TraceArray() As String = traceA.Split(Chr(44))
     For Each s In TraceArray
          WaveformGraph1.PlotYAppend(Double.Parse(s))
     Next
     mySession.Dispose()
Is this somewhat correct?  Any help is greatly appreciated.  Thanks!
Aaron
(3,557 Views)
Hi Aaron,
You might want to try an example program to see if you can get it up and running.  You can find them in C:\Program Files\National Instruments\MeasurementStudioVS2005\DotNET\Examples\VISA.
As for the asynchronous methods, there is a KnowledgeBase that describes this in great detail and may be helpful to you.
Hope this helps!
Adam W
Applications Engineering
(3,523 Views)
Your program has a very simple problem:  Note the asterisk in line 4.
     TextBox1.Text = mySession.Query("*IDN?\n")
     mySession.Dispose()
End Sub
The Agilent spectrum analyzers give the "query unterminated" error any time a particular query is not understood.
Good luck,
Joe
(3,495 Views)
