NI4882 equivalent command for the IVI formatted4​88IO ReadSTB command (AQ6317B instrument)
I cannot find in the AQ6317B optical spectrum analyzer manual the GPIB command that reads the status byte. The service command SRQ? only returns 0 or 1 meaning that the service is ON or OFF. Also SRMSK does not help. I need to read the status byte using (a) GPIB command(s) in order to find out when a longer operations such as sweep is done.
Can anyone help?
Thank you,
Cristian
(5,044 Views)
After researching this problem I have found that I should use NI4882.Device.SerialPoll and read the flag returned by this operation. However, in the latest version of NationalInstruments.NI4882.dll, v13.0.45.159, which I got a week ago from the NI website, this enum has the following values:
    [Flags]
    public enum SerialPollFlags
        // Summary:
        //     Device that is requesting service has a message available.
        MessageAvailable = 16,
        // Summary:
        //     At least one bit in the Event Status Register (ESR) is set.
        EventStatusRegister = 32,
        // Summary:
        //     Device that asserted the service request is requesting service.
        RequestingService = 64,
which as you can see, does not contain the values that I need: 1 !!!! It starts at 16, meaning b4 (the fourth least significant bit), and it's missing all the other values 1, 2, 4 and 8.
Anyone hit this problem? It seems very strange to miss these values. I need the value 1 because it is the bit that is set when an long operation like sweep finishes.
Thanks in advance,
Cristian
(5,038 Views)
To read the status byte, you use the STB? query. That's not the same thing as the ESR.
(5,030 Views)
Using STB? with my NI MAX, I get this answer to my sweep  '1250.00,1000.0,-30.0,  5.0, 2.0, 1,2501,NR_HLD,SGL,MEAS'. I was expecting a string representation of a byte, as the name of the command reads. My selected wavelenghts range is 1549 to 1551, 1000 measuring points. From this array of strings, I can only sassume that the 1 mans that the sweep is done and experiment from here on.
However, in the mean time I found another method. If anyone is interested: use NI4882.Device.SerialPoll() and read the result. The least significant bit (b0) of the result byte should be set.
Since I found the solution, I consider this matter closed.
Thank you anyway,
Cristian
(4,985 Views)