Hi Paul,
As you noted, the signal you have lacks a porch. That is the area that the 1410 uses to get a ground reference - the video signal is ac coupled, so the 1410 needs a reference to know where black is. This is known as the clamping region. Because your signal isn't flat in the clamping region, the DC restore circuitry starts oscillating, and you get the line pattern that you see.
In MAX, there are two parameters, called clamp start and clamp stop (I think - I don't have MAX in front of me). If you decrease those (it may take some fiddling), you should be able to get the 1410 to clamp on the sync tip instead. Those values are the number of pixel clocks to the start of the clamping area. Since you have a scope, you should be able to calculate the clamp start and stop values to use.
Once you have the 1410 clamping on the blanking area, you will get an image without lines. The contrast, however, will be way off.The black and white level settings are referenced to the clamping level (that was the whole point), but the clamping level was just dropped down about 300mV. If you adjust your black and white levels to compensate, you should get a good image.
Good Luck!
Jeff