Dear all,
I am generating a sine at 1 kHz (200 mV p-p) from the DAC NI 9263 (through FPGA, NI 9111 and NI 9012) and I am measuring the spectral purity at the output (see attached figure).
I see that the 2nd and 3rd harmonic are 50 -60 dB lower than the 1st harmonic whereas for my application it should be lower than 70 dB .
The spectral purity of the digital sine sent to the DAC is not the limit since the harmonics are lower than 100 dB .
Does anybody know the non-linearity of the NI 9263 DAC?
Has anybody generated a high spectral purity sine from a NI 9263?
Regards
Max
06-11-2012 10:52 AM
Hi,
Only one question...
How many samples are you generating in one period of a sine?
Could you increase the number of this value and observer the difference in amplitude from the harmonics?
Regards,
Claudio
06-12-2012 03:11 PM
Hi,
we have 96 points in the sinusoid generated from a Look up Table , and when we calculate the power spectral density the 2nd and 3rd harmonic are very low. We will try with 960 ponts to see the behaviour
Regards
Massimo