nimxs terminates at system startup after LV8.. installation and how I fixed it

Puhhh,
First I just wanted to (only Piscadela do Smiley ) upgrade from LV 7.1 to 8.0 ... 8.0.1  on my W2K system ...
Always after startup (restart) an Visual C++ error  (Smiley surpreso) pops up indicating that nimxs.exe produced an error and the process was terminated.  Result: No devices in MAX.Smiley triste
 
The hotline and some postings in this forum indicated that an un and reinstall of (all?) NI Software might solve the problem.  Oh oh, anotherone day bites the dust  (FREDDY  HELP ME!)
 
Since all our office PCs have to switch to XP this year I decided to start from scratch ...(next day)...
However, I still got this Error message. Smiley maluco
 
Another research at the entire NI site and the hint about MAX 4 that helped: http://digital.ni.com/public.nsf/allkb/50600969c3bfefe086257067005125cd
 
with a minor variance :
 
1. exit all NI software
2. rename the Data directory at C:\Program Files\National Instruments\MAX\Data  to Data2
3. reboot
 
Smiley muito felizSmiley muito felizSmiley muito feliz
 
the only thing that is missing now is the mass compile      .... and I have to change my signature
 
 
Greetings from Germany
Henrik

LV since v3.1