Hi Craig,
Can you explain in a little more detail your test setup? In particular, how are you obtaining the records/second benchmark?
With the trigger setup the way you describe, every 500uS it will
receive a trigger. When this trigger is received it will take 1 record,
with a minimum record length of 500 samples.
The total records to be acquired is 2000.
I setup a PXI-5105 triggering off a 2khz square wave on PFI1, the
digitizer is set to acquire 2000 records at 500 samples per record. I
timed this and it's coming in around 750mS every time. I'm attaching
screenshots so you can see the setup. I've attached the example I used
to this post. You will need to configure the Resource name (DAQ card),
Sample Rate, Number of samples per record (minimum record length) and
number of records. On the block diagram specify the Trigger source. The time readout is in mS.
The PXI-5105 should not miss a trigger if it's configured correctly.
Have a great weekend!
MatthewW
Applications Engineer
National Instruments
Message Edited by Matthew W on 08-03-2007 03:01 PM