Hi Mike,
Gabi’s suggestions are definitely good ideas. If you are certain that your Speedmaster device is generating a TTL signal, you can check to see if your signals are connected properly using MAX.
1.
-Open Measurement and Automation Explorer.
-Right-click on the 6024E and select “Test Panels…” Test Panels are useful for trouble shooting because it removes any LabVIEW code from the test. You can use the digital I/O tab to see if you are switching the digital value. You can use the counter I/O tab to see how many pulses are counted.
-To configure the Counter I/O test, under Channel Name, select whichever counter you are using (zero or one).
-Change the Mode to “edge counting”.
-Select your signal channel as the “Edge Source”. The signal channel for your device (assuming you are using counter 0) should be PFI 8.
2.
There are several examples for acquiring frequencies in LabVIEW. These are accessible in LabVIEW by selecting Help>>Find Examples…
Then browse to Hardware Input and Output>> DAQmx >> Counter Measurements >> Digital Frequency. The appropriate example for your application would be “Meas Dig Frequency-Low Freq 1 Ctr.vi”.
3.
For simple acquisition tasks, it may be easier to generate your own program using the DAQ Assistant.
-Place a DAQ Assistant on the Block Diagram and choose Acquire Signals>>Counter Input>> Frequency.
-Select your device and counter.
-Configure your Signal Input Range. You mentioned that your input will be around 8000 Hz.
-Your Measurement Method should be “1 Counter (Low Frequency)”.
-For acquisition mode, you will probably want to use either “N samples” or “Continuous”.
-“Samples to Read” will be determined by how many pulses of which you want to measure the frequency. You may want to measure the frequency of a large number of pulses and average them to get one value.
-Click “OK”.
-On data output terminal of the DAQ Assistant Express VI, right-click and select Create>>Indicator to create a graph of the measured frequency.