Hi - Alan from LeCroy Tech Support here.
Measurement and Automation Explorer is designed to include instruments that are VXI-11 or LXI compliant. The WavePro 940 is not, so you cannot add it to MAX in this way. Please note that there is no need to include the instrument within MAX in order to use it in LabVIEW.
Feel free to contact me if you need further assistance!
Best Regards,
Alan
07-16-2007 02:59 AM
Dear Alan,
Thanks for your help!
I tried to connect Labview directly with the Wavepro 940, without using MAX. Within the Functions Palette of Labview I can select from the lcwpxxx-driver. So I guess the Wavepro driver is installed correctly. But when I open a Front Panel and give as Resource Name "TCPIP::<my IP-address>::1861::SOCKET", nothing happens. In this way it should work, no? Or do I do something wrong here?
Best regards & thanks in advance!!
Floris
07-23-2007 02:21 AM
When you control a network instrument through TCPIP SOCKET interface, every device command must explicitly adds the "termination character" such as line-feed (0x0A). For example, your must send "*IDN?\n" rathr than "*IDN?".