Want more details of how cRIO scan mode works with NI-9237
The NI-9237 is an analog acquisition C-series module for strain gages and load cells. I'm using it in a NI-9144 EtherCAT backplane connected to a cRIO-9024 (on a NI-9114 backplane). It's used with four 5000Lb load cells.  While the system runs in hybrid mode (there are other module interfaces that only work in FPGA mode), I hope to run the NI-9237 module in scan mode, but there is an "offset calibration" feature/property of the module that I would like to use that is not accessible under scan mode. The Offset Cal property is accessible to FPGA code. My question is the following..
Does the NI-9237 scan mode interface turn on and take advantage of the offset calibration/correction feature within the FPGA portion of the "RIO Scan Interface" code or does it leave it ignored and turned off?    I cannot find any document that says that scan mode does use it but I also cannot find any document or person who can say that scan mode does not use it.   As it removes instrumentation offsets from the measurement, it seems like something that would be good to have on so it seems reasonable that the scan interface might naturally use it.
Setting Auto Zero on a NI-9237 with NI-RIO
The RIO Scan Interface Under the Hood
Even though the FPGA and input module hardware can in most cases acquire readings considerably faster than the scan interface can accept them, allowing for the possibility of oversampling with averaging or filtering before delivery of a measurement to the scan interface, the "Under the Hood" reference above seems to imply that only individual measurements are made and delivered by the FPGA to the scan engine. Is this implication correct?
“Each module performs as many conversions as possible to provide the most recent I/O values to I/O variables configured for direct I/O access, which bypasses the scan and reads directly from hardware.”
(3,778 Views)
Accepted by topic author WNM
Hi WNM,
You are correct in that the NI-9237 does not support that property in scan-mode.  When it is being used in scan mode, the offset calibration/correction feature is not turned on.
The scan engine operates using I/O variables without any buffering.  It reads the most recent values stored value in the variable one at a time.  For more details, please see this help document and reference the Scanned Access section: http://zone.ni.com/reference/en-XX/help/371361H-01/lvioscanhelp/io_vars/#scanned .
Hope this helps!  Happy New Year!
Dayna P.
(3,734 Views)
Dayna,
Thank you for the reply. Prior to ever posting this question to the forums I had first ask it through our official NI support channel but since that answer was taking forever to arrive, what with the end-of-year holidays and all, I decided to try the forum. Your answer beat theirs by about a week but in the end they both were the same. Thanks again!
(3,667 Views)
