Using HV application with DAQ
I recently built a portable high voltage, low current adjustable power supply for a project I have been working on the last couple months. The device simply takes an incoming dc voltage, uses a EMCO DC to HV DC converter, and outputs the high voltage. Also, the high voltage is also reduced back by 1000 using a EMCO voltage divider, so the user can safely put a voltmeter to measure the output as they vary it for different applications.
I am also putting together a vi that will allow them to automatically log voltage and current and allow them to store the data in a spreadsheet and view a graph. This part is done, but I am having one problem interfacing with DAQ and need some help or advice on how to make it work, if possible.
For example: If i measure the output of the voltage divider (+volts and neg common), the output would be 5Vdc (which translates to a 5KVdc output on the power supply using a Fluke multimeter). However, when I attach my USB 6008 DAQ I/O to it, then the voltage will drop to 1.1Vdc. I am using a differential config in MAX because it may range to 12V on some models I built. My question is why is DAQ dropping the total output voltage?
Thanks in advance for any help or advice.
(3,403 Views)
The input impedance of the analog input of the 6008 is 144Kohms. I suspect that it is loading down or changing the ratio of the EMCO voltage divider. You may need to add a unity gain buffer between the EMCO divider and the 6008 so as to not load it down. A Fluke meter typically has in input inpedance of 10Mohm or greater, so it will not load the divider doen.
-AK2DM
(3,389 Views)
Makes sense.....I will give it a try and let you know if it works or not. thanks!
(3,374 Views)
I had posted as perrin.mr:
to the person who suggested impedance was the issue: You were correct. Thanks!
I ended up using a 6210 DAQ (input impedance of 10Gohms) instead of making a unity gain buffer to use with the 6008. After some minor scaling in the program my reading are about .03 tolerance. That is accurate enough for my application.
Thanks again
Jody M.
The Procter and Gamble Co.
Currently using LV2018.
(3,346 Views)
Very good, I don't know why NI did not put a better front end on the 600x's, probably a cost saving measure.
-AK2DM
(3,340 Views)
