Dear all,
when I'm trying to read measurement values from thermocouples and pressure sensors (voltage input signal), the pressure values would be 1 to 1.2 seconds delayed. The thermocouples, however, are responding to changes immediately.
Does it have to do with the two VIs for starting a task that I put in series (This is necessary as I can't define a conversion including an exponential function)? The sample rate was set to 10/s.
Best
Sebastian
Message 1 of 10
(3,399 Views)
Hi Sebastian,
you can simplify the expression node like this:
(The LabVIEW help explains how to calculate the power of numbers!)
When you are talking about those DAQmxCreateVirtualChannel functions: no, it doesn't matter how many channels you configure before the DAQmxStart function. (Atleast not for a delay of >1s.)
Message 2 of 10
(3,364 Views)
Thanks GerdW,
I'm not quite sure how I can simplify the node. But anyway, even removing the note does not eliminate the time delay.
The time delay is, however, not observable if I remove the thermocouple acquisition. Any idea on that?
Best,
Sebastian
Message 3 of 10
(3,296 Views)
Hi Sebastian,
thermocouples are (usually) read at a very low sampling rate.
Can you verify/check that you are reading them at the expected sampling rate?
Your NI9214 only supports 68S/s aggregated sample rate! And that DAQmx task is configured containing 3 modules with 40 thermocouples…
Message 4 of 10
(3,291 Views)
Thanks GerdW,
does an aggregated sample rate of 68 S/s mean that I can acquire 68 Signals per second for the whole module? So in case of 10 thermocouples per module, 6.8 signals per second maximum? And in case of 40 TCs in 3 modules 204/40 = 5.1 signals per second?
Best,
Sebastian
Message 5 of 10
(3,282 Views)
Hi Sebastian,
Yes. And with 16 TC connected to your module you only get 4.25 S/s/ch!
The module is able to read 68 samples per second. With just one channel you get 68S/s/ch, with 7 channels you are down to ~10S/s/ch…
I'm not sure how DAQmx handles the 3 modules in the task.
But I guess it will be limited by the slowest module in the task, so you should get 68/16 = 4.25 S/s.
(You may distribute your TCs on your modules like 13+13+14 to yield an effective sample rate of 68/14 = 4.86S/s - an improvement of 14%…)
Message 6 of 10
(3,274 Views)
Hi Gerd,
this is such a good help, thank you Gerd.
Indeed, if I reduce the number of thermocouples to two, there's no delay in the pressure readings anymore.
Unfortunately it does not work to to set the thermocouple sample rate to 1/s and the pressure sensor sample rate to 10/s - in this case there's the delay again.
Best,
Sebastian
Message 7 of 10
(3,259 Views)
Hi Sebastian,
you can handle both devices/DAQmx tasks in separate loops. Now you could sample as fast as possible/needed for TC and pressure data…
Message 8 of 10
(3,241 Views)
Hi Gerd,
 according to your suggestion I set up one loop for temperature value acquisition running with 1 Hz and a seperate one for pressure value acquisition running at 10 Hz. This still gave me a delay of about 1 s in the pressure values until I triggered both tasks to start reading at the same time, which finally solved the problem.
Thanks again and best regards,
Sebastian
Message 9 of 10
(3,208 Views)
Hi Sebastian,
glad you found a solution.
Be aware of possible problems when running your VI for long-time measurements! As the modules use different timebases their samplerates might differ from selected 1Hz/10Hz slightly (like 0.999Hz/10.002Hz), so you might notice small "delays" for long measurement campaigns.
Message 10 of 10